NFPA99C 5.1.3.6.1.2(5)
Piping between the vacum pump(s), discharge(s), receiver(s), and the vacum source shutoff valve shall be in accordance with 5.1.10.2 except that stainless, galvanized, or black steel pipe shall be permitted to be used.
The plumber wants to use galvanized pipe on the discharge lines. The manufacture requires non-corrosive pipe, the engineers, designer, installer, and the hospital are all okay with the galvanize pipe but want my approval (local AHJ)
Anybody have pros or cons comments
Piping between the vacum pump(s), discharge(s), receiver(s), and the vacum source shutoff valve shall be in accordance with 5.1.10.2 except that stainless, galvanized, or black steel pipe shall be permitted to be used.
The plumber wants to use galvanized pipe on the discharge lines. The manufacture requires non-corrosive pipe, the engineers, designer, installer, and the hospital are all okay with the galvanize pipe but want my approval (local AHJ)
Anybody have pros or cons comments